Chlorine experiment

Ian Wilde
Administrator
This post was updated on .
I'm running a chlorine experiment for our community patient setup so thought I would post the whole thing here for comment and scrutiny.

Kit:

We're using a single patient RO.  The DWA one is what we have in around 50% of our homes, the other 50% use the Aqua Uno which is mechanically identical.  

We'll be taking the measurments with our Chlorosense meter. Our Chlorosense test strips are stored in the fridge and the calibration of the meter is to be verified before every test using the calibration standards supplied by Palintest.



The tap water is straight from the main and after a quick measurement this morning we recorded an incoming chlorine level of 0.8ppm.

This water is going to feed through a 10 micron particulate filter which is connected in series to another filter bowl which will house our current carbon filter - the Pentek C1.



To ensure a realistic contact time between carbon and water we are going to feed the RO permeate through a regulator set to give a flow of approximately 500mls/min.  This should simulate an average draw from a machine.  The drain from the RO will obviously be acting as normal with a flow of approx 2000mls/min.

We will measure daily to see how the carbon begins to exhaust making note of how many running hours/litres have occurred.
Measurements will all be taken at least 1 hour after starting the RO every day to ensure accurate contact time.

We will plot 3 axis on our chlorine graph; incoming water, post carbon and just out of curiosity, post RO.

If you have any comments as the experiment unfolds then please feel free to post within the thread.

Thanks.


Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Todays readings - just for info...



The chlorosense shows a reading of <0.05 total chlorine.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Ian Wilde


So far we have clocked up 15 hours worth of water passing through the filter which equates to 3 dialysis sessions at 4 hours each with a disinfect of the machine included (ish).

For our home patients the filter needs to last around another 45 or so hours with our current filter changing regime.


To be continued on Monday...
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Gareth Murcutt
Hi Ian,

Did very similar experiment a few years ago. We just left it running at 1l/min and did >50,000hrs. I can email it to you or is it possible to attach/post docs on here.
Cheers
Gareth
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Chris Bates
Unless the test is still running 7 years after I left, I suspect you mean litres not hours !!
Yours
Peter Pedant

p.s If  I remember that test - it was on Matrikx 10" carbon blocks which completely blew the impregnated filters away.
The latest CFB impregnated type now seem to outperform blocks by quite a way.
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Gareth Murcutt
Hi Gareth,

If you've anything that is an image you can certainly insert it into posts otherwise I think you can only copy and paste.
If you want to send it to me anyway I can always take a few screen shots and we can compare experiments
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Ian Wilde
Today's readings:



I have to admit that it is doing better than I thought but as well as I hoped - if that makes sense!
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Ian Wilde
Some graphical info...

Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Today we breached at Post Carbon but not Post RO.  Incoming chlorine levels were higher today than any other day too.





Will keep going until the water that the patient is exposed to (RO Permeate) hits the limit.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Morgan
Interesting Ian,
So if you passed about 3500 litres, this would be about 7 sessions of 5 hours, including the disinfect at the start and end of each session. Were you expecting more from the filter.
I suppose the next thing is to try testing a few carbon block filter, that will hopefully give longer running time. This also goes to show how having redundancy is imprtant as the chlorine went high very quickly.

Ian
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Thanks for the comment Ian,

The filter hasn't performed as well as required tbh. Our filter changing regime is once a month and this experiment has shown that the filter alone has only lasted around two and a half weeks of a thrice weekly dialysis program.

From a safety point of view the patient is still being exposed to chlorine levels well below the allowed limits - phew!  
Should the RO be considered part of the chlorine removing setup?  I think not but it can provide the redundancy that you mention I suppose!?

I think I am going to continue the experiment with the current filter to see how the exhaustion pans out and see just how far we can push it before a patient risks exposure with the protection of the RO.  I think it would be interesting to connect the RO straight to tap water and see what results it brings.

I have a carbon block filter standing by for stage 2 so we'll hopefully see a dramatic improvement.

If anyone out there knows how to calculate the approximate contact time using a cartridge filter and the flows I've been tabulating I think it would be interesting to see what it is.  
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Chris Bates
Hi Ian
The manufacturer specifies capacity of the 10" C1 is 11250 litres at 3.8 l/min flow. Capacity tested at given flow
rate using 2 ppm free available chlorine to 0.5 ppm breakthrough. So this is a bit disappointing.
I think it would be worth testing the latest CFB-plus filters. Each 20" CRB filter is specified at 20 times the capacity you measured.
It would save me putting any effort in to testing it myself too !
Chris
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Cheers for the info Chris, I guess manufacturer's specs and reality can be two different things!

Our current filter setup is all in 10" bowls so my next test is with a 10" carbon block to see if we can get away without having to mod our installs.

Depending on our next test it could influence whether we upgrade to 20" bowls for future installs at least.

Either way I know our patients are safe on the monthly changing regime but my goal is to change just every 6 months when the techs visit to take samples etc... I know the patients would love to lose the hassle of changing them.

For me to put it forward though at our next MDT I'm going to have to have the evidence to back it up.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

fraser gilmour
Hi Ian, email me your spread sheet and I can input the test data from one of our units if you like. It's a satellite unit with 10 stations where we have a small central plant using chlorplus block cartridges for chloramine removal. Might be interesting to compare?
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
Hi Fraser,
Will defo do that when I'm back in tomorrow.
Been out of the workshop for most of the day.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Ian Wilde
We're getting there...





I make that roughly 11 dialysis sessions at 4 hours with disinfects.
Almost 3.5 weeks.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
In reply to this post by Ian Wilde
Well I think we'll end Phase 1 here:





Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

Ian Wilde
Administrator
After sample 9 not only had we broken through the 0.1ppm limit at the output of the carbon filter we had also started to lose any redundancy that the RO was giving us.

We had processed ~4699 litres of water which equates to roughly 465 litres per 5 hours of machine operation (dialysis/disinfection)

4699/465 = 10.1 dialysis sessions.

This is roughly 3.5 weeks so our monthly changing regime is just about hanging in there.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

Re: Chlorine experiment

fraser gilmour


The data is from a plant supplying 10 stations @ 3 shifts per day from 19/9/14 to 26/2/15. The filters used are x2 chlorplus 20" and x2 chlorplus 20"BB. Filters were changed after sample eight.
The data shows that over three months and approximately 1500 cubic meters of water passed, the filters are able to reduce the chlorine level down below the limit for feeding into the RO, even when our water is at it's coldest and the chloramine level is highest.It also shows that we have a decent amount of spare capacity if the level were to unexpectedly increase at the end of their life, even without relying on the RO to reject.
It's possible that we could make some efficiency savings here so I'm going to experiment with a different filter configuration over the next year.
12