Hi David, thanks for the response. Braun are currently looking into the issue for us. We currently run at 800ml/min dialysate flow for HDF and this may turn out to be contributing to the prefilter pressure hitting the alarm point. Unless you also run at that flow rate without issue? I’d heard the filters could run way past the recommendation, but as per my post, falling around a third short. We’d only follow the manufacturers guidance but good information to have
.
We only have 24 IQs. I’m aware of a few unexplained bloodside test fails though most of these have passed on subsequent checks. On the few times I’ve seen the issue as it is happening, adjusting the pods seems to have got it through, though may have been coincidence. Again, only in use for just over 6 months.
Have had two mSATA fails, one after around 300 hours the other at just over 2000 hours.
|