Just wondering how people interpret the following:
4. Operation and maintenance of water treatment systems for haemodialysis
Guideline 4.1 – Routine maintenance and monitoring
Guideline 4.3.2 – Frequency of monitoring
The frequency of monitoring is generally recommended by the equipment manufacturer or by local operating procedures. We recommend that the minimum frequency of monitoring of water for dialysis should be as follows (1D): Contaminant Frequency of testing
Total chlorine At least weekly
Total viable counts At least monthly
Endotoxin At least monthly Chemical contaminants other than chlorine At least every 3 months
Frequency of monitoring of microbial contaminants and endotoxin
Total viable counts and endotoxin levels should be monitored at least monthly. A monitoring scheme should be drawn up which defines the points at which the sampling is performed. The sampling plan is dependent on whether a dead space free secondary loop is used to connect the dialysis machines to the ring main. If such a connection is used, then only one sample needs to be taken at the end of the loop.
Now, noted in
BS EN ISO 23500:2015 it states there is no requirement to test for bacterial growth or endotoxins when the haemodialysis system is fitted with endotoxin retentive filters that are operated according to the manufacturer’s instructions, unless the manufacturer requires such tests to be performed.
So if you are using Diasafes or Ultrafilters, ultrasafes etc... does that contradict the need for the sampling as per guideline 4.3.2?
If not does it still need to be monthly? Are we spending more than we need to on sampling?