Major air leakage, dark blood out of dialyser, FCN 5.11.001

classic Classic list List threaded Threaded
5 messages Options
Reply | Threaded
Open this post in threaded view
|

Major air leakage, dark blood out of dialyser, FCN 5.11.001

Paul Addison
Hi all,
This fault has occurred on 2 machines (AK200S) so far; several times at the start of dialysis, but also a couple of times approaching 2 hours into treatment. The blood from the dialyser to the bubble trap is noticed to be suddenly very dark, the venous pressure is high (220mmHg) and TMP rises. Then a 'Major air leak' message is displayed, later followed by error 5.11.001.
The machines seem fine in the workshop and I'm running out of ideas! Has anyone experienced anything similar before, or have any ideas as to the cause? We have some theories but nothing concrete yet.
Thanks
Paul
Reply | Threaded
Open this post in threaded view
|

Re: Major air leakage, dark blood out of dialyser, FCN 5.11.001

Ian Wilde
Administrator
Hi Paul,

Have you been doing dummy treatments in the workshop or just completing the function check and then monitoring?
Have you changed dialyzers at all?  Is the dialysate compartment being primed fully before the blood side?


Ian
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

RE: Major air leakage, dark blood out of dialyser, FCN 5.11.001

Paul Addison

Hi Ian,

I have performed simulated patient treatments but with no issues. It has occurred with fx80 and fx100 dialysers but we are keeping tabs on this. Interesting last question; here, they wait until the 4:00 is displayed before lining and priming and wait until it asks for the Bicart. At other units they line and prime when the priming light flashes. The 2 machines are now back in use as dialysers are currently the primary suspect.

Thanks

Paul

 

From: Ian Wilde [via ART Forum] [[hidden email]]
Sent: 08 July 2015 13:22
To: Addison Paul (ROYAL DEVON AND EXETER NHS FOUNDATION TRUST)
Subject: Re: Major air leakage, dark blood out of dialyser, FCN 5.11.001

 

Hi Paul,

Have you been doing dummy treatments in the workshop or just completing the function check and then monitoring?
Have you changed dialyzers at all?  Is the dialysate compartment being primed fully before the blood side?


Ian


If you reply to this email, your message will be added to the discussion below:

http://art-forum.1008460.n3.nabble.com/Major-air-leakage-dark-blood-out-of-dialyser-FCN-5-11-001-tp366p370.html

To unsubscribe from Major air leakage, dark blood out of dialyser, FCN 5.11.001, click here.
NAML


********************************************************************************************************************

This message may contain confidential information. If you are not the intended recipient please inform the
sender that you have received the message in error before deleting it.
Please do not disclose, copy or distribute information in this e-mail or take any action in reliance on its contents:
to do so is strictly prohibited and may be unlawful.

Thank you for your co-operation.

NHSmail is the secure email and directory service available for all NHS staff in England and Scotland
NHSmail is approved for exchanging patient data and other sensitive information with NHSmail and GSi recipients
NHSmail provides an email address for your career in the NHS and can be accessed anywhere

********************************************************************************************************************
Reply | Threaded
Open this post in threaded view
|

RE: Major air leakage, dark blood out of dialyser, FCN 5.11.001

Ian Wilde
Administrator
Well, yes you certainly can prime once the light is flashing but here we tend to wait until the 4 hours is up on screen too.
Ian Wilde
Reply | Threaded
Open this post in threaded view
|

RE: Major air leakage, dark blood out of dialyser, FCN 5.11.001

Paul Addison
It has now occurred on 2 further machines so it seems more likely to be down to something other than particular machines, though it's still a mystery