NIKKISO DBB 05.Conductivity dropping

Maxence11
Hello all,
I have an issue on a DBB-05 machine. Preparation of dialysate is ok. But when we start dialysis, the conductivity drops. Bicart cond is around 3.0 but total conductivity is dropping during dialysis.
Can someone help me.
Reply | Threaded
Open this post in threaded view
|

Re: NIKKISO DBB 05.Conductivity dropping

John Moore
I assume you have rebuilt the acid pump with new seals?
Reply | Threaded
Open this post in threaded view
|

Re: NIKKISO DBB 05.Conductivity dropping

Maxence11
Yes. Now it's ok.
Thank you very much