UF Test Failed GAMBRO AK200S

classic Classic list List threaded Threaded
3 messages Options
Reply | Threaded
Open this post in threaded view
|

UF Test Failed GAMBRO AK200S

Maxence11
Hello all,
I have a problem on GAMBRO AK200S. During dialysate preparation tests, the machine passed too much time on leakage test.And after, it shows that the UF AUTOTEST FAILED  and it will restart in 5min.
Can someone help me please
Reply | Threaded
Open this post in threaded view
|

Re: UF Test Failed GAMBRO AK200S

Ivor D
I'd check for any potential leakages between the FOP and FIP. If there are no obvious issues, an FRN calibration could well solve your issue.
Reply | Threaded
Open this post in threaded view
|

Re: UF Test Failed GAMBRO AK200S

Maxence11
I have calibrated the FRN and everything is ok now.
Thank you very much!