ART Forum
Search
everywhere
only in this topic
Advanced Search
UF Test Failed GAMBRO AK200S
Classic
List
Threaded
♦
♦
Locked
3 messages
Maxence11
Reply
|
Threaded
Open this post in threaded view
♦
♦
|
UF Test Failed GAMBRO AK200S
Hello all,
I have a problem on GAMBRO AK200S. During dialysate preparation tests, the machine passed too much time on leakage test.And after, it shows that the UF AUTOTEST FAILED and it will restart in 5min.
Can someone help me please
Ivor D
Reply
|
Threaded
Open this post in threaded view
♦
♦
|
Re: UF Test Failed GAMBRO AK200S
I'd check for any potential leakages between the FOP and FIP. If there are no obvious issues, an FRN calibration could well solve your issue.
Maxence11
Reply
|
Threaded
Open this post in threaded view
♦
♦
|
Re: UF Test Failed GAMBRO AK200S
I have calibrated the FRN and everything is ok now.
Thank you very much!