We test our home patient machines twice a year for TVC and Endotoxins and once a year we carry out a RAT1 Chem check, this all happens at service time. Could someone please enlighten me, is there a guideline specific to home install testing as I'm going through a phase of updating procedures and would like the correct info
Thanks John
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: home water test

Ian Wilde
Administrator
612 posts
Hi John,  
We do the same as you. We chose this based on the recommendations in the RA guidelines and also considered the consistent quality of feed water from the utility company. Our water results trend data over the last X years also supports this decision. As with any test it's only as good as the moment it is taken so personally I feel trend data becomes more important.
Ian Wilde
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: home water test

Chris Bates
231 posts
Ditto
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: home water test

mike.jones
60 posts
We are in the process of moving from 4 to 6 months following consistant results.