Hello,
I have a Gambro AK200 machine that can not finish disinfection..is shows me -01. I can remark that the degassing pump is not running and also the blood leak trap is empty. I have replaced the degassing pump and the DRVA. Also I have replaced the recirculation valve. But no changes.
Can you please advise me some ideas..
Does the degas pump run in Fch? and draw round fluid? If not it maybe Fm CPU brd. Have you checked the stirrer motor? As it could drag the degas pump down, although I would expect this to spanner during Fch.
Drew
Drew Blackmore
Senior Renal Tech
Royal Free London