It took over a week of patients coming off heavy for staff to tell us about this machine, and they only casually mentioned it when they reported the machine for a separate problem
When I spoke to staff about it they said they would only report a relatively large UF discrepancy, far more than the machine + weighing scale tolerance so I don't think you can rely on this fault being picked up with the regular pre/post weighing.
It's probably low incidence of a low risk fault, but I think it's important to highlight that it can happen as I didn't believe it could. I assumed the integrity of the pump was fully checked in T1 and it isn't.
I think the only way to pick it up is by a simulated treatment UF check